In this clinical investigation, including 248 adult patients with a wide range of conventional laryngoscopic views requiring orotracheal intubation for elective thyroid surgery, tracheal intubation with the Airtraq Optical Laryngoscope (AOL) performed by novice but instructed users was successful at the first attempt in 97 percent with successful tracheal intubation after a maximum of two attempts in all patients. These favorable results may not be obtained when operators without formal instruction in the technique perform tracheal intubation attempts with the AOL.

 

The results of our study strongly support the initial use of a suitable airway model for formal hands-on training; this approach was obviously the key factor for success. When the novice users of the AOL performed the first series of laryngoscopy and tracheal intubation attempts on patients, they were already familiar and confident with the technique. The number for obtaining acceptable clinical performance and competence with this device, estimated by the instructor, remained within 10 applications for all operators. The clip-on AOL Video System instead of the removable rubber coat on the view finder allows viewing of the entire procedure on a compatible medical monitor, not used in this study; simultaneous viewing by the operator and the instructor might have even accelerated the instruction.
 

The relatively bulky AOL blade, a limitation of this device, may impede blade insertion, particularly in patients with limited mouth opening and/or restricted pharyngeal space (encountered in 10 percent). However, the AOL blade favorably incorportes the tracheal tube (TT) already in the tube-guide channel, eliminating the need for additional space passing the TT through the oral cavity.  

 
The view finder of the AOL allowed visualization of anatomical structures with adequate image quality and provided a panoramic view of the laryngeal aperture when the blade was properly inserted. There were no problems with poor visibility due to fogging of the optical system by warming-up the system slightly above body temperature within a short period of time, a special feature of the AOL; also secretions did not appear to interfere with the laryngeal views because the optical system is protected and remote from the tip of the blade.
 
Visualization of the entire laryngeal aperture, a prerequisite of successful AOL-assisted tracheal intubation, failed in only one patient at the first attempt but was finally obtained at the second attempt. The curvature of the tube-guide channel and the orientation of the optical system of the AOL were usually well aligned with the laryngeal aperture; when the laryngeal aperture was in the center of the view finder, advancement of the TT into a midtracheal position was generally successful at the first attempt, except in 6 patients. Nevertheless, there were minor problems and difficulties with impeded TT passage through the laryngeal aperture in 10 percent, partly attributable to distortion of the larynx, frequently observed in patients undergoing thyroid surgery. When the TT impinged on laryngeal structures, usually on the arytenoid cartilages or the interarytenoid area, a backward, upward, and to the left or right movement of the AOL blade regularly solved this problem. Interestingly, impeded TT advancement during tracheal passage despite tracheal deviations and/or compressions by enlarged thyroid glands and using standard TTs was rarely a problem.

 

Severe complaints and injuries associated with airway management were not observed; this may be due to vizualization of the entire laryngeal aperture during successful laryngoscopy and controlled TT advancement into a midtracheal position. Conventional tracheal intubation, particularly after multiple attempts, may not provide comparable results. It should be noted, that the insertion of the AOL blade into the oral cavity was performed with special caution because this maneuver is usually blind; pharyngeal injuries by uncontrolled blade insertion cannot be excluded.


The efficacy of the AOL in difficult CL situations is now supported also by other clinical studies. Maharaj and co-workers (3) compared the ease of tracheal intubation with the AOL or the Macintosh laryngoscope (ML) in patients with normal airways but with manual cervical spine immobilization in a randomized, controlled study (with 20 patients in both groups). All patients, except one patient in the ML group requiring three attempts, were successfully intubated at the first attempt; the AOL reduced the duration of the intubation time and the need for additional airway maneuvers. Furthermore, they compared the success of tracheal intubation with the AOL or the ML in patients with anticipated difficult CL (again with 20 patients in both groups) (4). AOL-assisted tracheal intubation was successful in all patients after a maximum of two attempts (in 19 patients at the first attempt); even after three attempts, conventional tracheal intubation failed in 20 percent, but these 4 patients were successfully intubated with the AOL at the first attempt. Favorable results were also published by Ndoko and co-workers (11), comparing the success rate of tracheal intubation with the AOL or the ML at the first attempt in morbidly obese patients (with 53 patients in both groups). Tracheal intubation with the AOL and the ML was successful in 100 and 89 percent, respectively; the 6 patients with unsuccessful ML-assisted tracheal intubation were successfully intubated with the AOL at the first attempt. Obese patients were given preferentional enrollment into our study; we also had less problems with successful tracheal intubation in this patient population.  

 

In a case series of 7 patients with difficult CL (grade 5, visualization of only the soft palate, despite external laryngeal manipulation) and failed tracheal intubation after multiple attempts with CL, Maharaj and co-workers (7) were sussessful at the first attempt of AOL-assisted tracheal intubation with views of the entire laryngeal aperture. We agree with their suggestion, that the AOL may serve as valuable rescue device in situations of difficult CL and failed tracheal intubation.

 
Provided formal instruction, the success of AOL-assisted orotracheal intubation performed by novice users was not affected by conventional laryngoscopic view grading. The AOL proved to be uniquely useful for routine and difficult laryngoscopy and tracheal intubation in clinical practice. The AOL may be also useful in medical emergency situations, but this suggestion has to be supported by further studies.

 